A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)
Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.
Correct answer is '90.90'. Can you explain this answer?
Verified Answer
A thermopile having a resitance of 100 Q and consisting of 20 copper-...
The thermocouple output is directly proportional to the temperature difference between hot and cold junction. Junction RJ is at a temperature of 25°C.
The internal resistance of voltmeter of resistance 1000Ω having a voltage of 1.47 mV
View all questions of this test
Most Upvoted Answer
A thermopile having a resitance of 100 Q and consisting of 20 copper-...
The thermocouple output is directly proportional to the temperature difference between hot and cold junction. Junction RJ is at a temperature of 25°C.
The internal resistance of voltmeter of resistance 1000Ω having a voltage of 1.47 mV
Explore Courses for GATE exam

Similar GATE Doubts

A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer?
Question Description
A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? for GATE 2024 is part of GATE preparation. The Question and answers have been prepared according to the GATE exam syllabus. Information about A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? covers all topics & solutions for GATE 2024 Exam. Find important definitions, questions, meanings, examples, exercises and tests below for A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer?.
Solutions for A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? in English & in Hindi are available as part of our courses for GATE. Download more important topics, notes, lectures and mock test series for GATE Exam by signing up for free.
Here you can find the meaning of A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? defined & explained in the simplest way possible. Besides giving the explanation of A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer?, a detailed solution for A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? has been provided alongside types of A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? theory, EduRev gives you an ample number of questions to practice A thermopile having a resitance of 100 Q and consisting of 20 copper-constantan thermocouples is used to measure temperature difference between two points. The temperature of the first point measured separately is 25°C. EMF generated measured by a voltage measuring device having an internal resistance of 1000 Q is 1.47 mV. E MF-temperature relationship for copper-constantan thermocouple with reference junction at 0° is as follows. (2001)Determine the temperature difference between the two points after applying a correction for loading effect in the measurement of the EMF.Correct answer is '90.90'. Can you explain this answer? tests, examples and also practice GATE tests.
Explore Courses for GATE exam
Signup for Free!
Signup to see your scores go up within 7 days! Learn & Practice with 1000+ FREE Notes, Videos & Tests.
10M+ students study on EduRev